NickV Posted June 2, 2017 Report Share Posted June 2, 2017 I've bought some 2500kg submersible load cells which I'm connecting now to my Vision 430 J T2 PLC. I have a LC3 expansion module. Since it is almost impossible for us to calibrate the load cell according to the help files (putting real weight) on it, I'd like to use the calibration report the supplier gave me and use the 'Edit calibration point' function to match raw values with weights. Can someone give me insight on the way I have to calculate my calibration points based upon the report of the supplier? When I connect the load cell and change the representation mode to raw value, I see a raw value of around 60 000 at 0kg. 11942-2 Cal Cert.pdf Quote Link to comment Share on other sites More sharing options...
MVP 2023 Joe Tauser Posted June 12, 2017 MVP 2023 Report Share Posted June 12, 2017 First, are you using the load cell in AC or DC mode? I'd recommend AC. i have not done it this way but I can set up some hardware and fiddle with it if needed. The loadcell input is a 24 bit A/D, which translates to 16,777,216 counts. Try Mode 6, which is uV/V. The Help talks about mV/V but you have to have a way to enter the decimal places so it's really uV/V. What do you get when you use this mode? Post your code, too. Joe T. Quote Link to comment Share on other sites More sharing options...
Gabriel Franco Posted June 12, 2017 Report Share Posted June 12, 2017 I´ve never done this before, but I guess Edit Calibration Point can be used for your purpose. It may be located on FB|Loadcell|Advanced|Calibration menu. Quote Link to comment Share on other sites More sharing options...